    for a world of water OceanOpportunity is a splendid one to behold OceanOpportunity is a splendid one to behold OceanOpportunity is a splendid one to behold OceanOpportunity is a splendid one to behold OceanOpportunity is a splendid one to behold OceanOpportunity is a splendid one to behold OceanOpportunity is a splendid one to behold OceanOpportunity is a splendid one to behold OceanOpportunity is a splendid one to behold OceanOpportunity is a splendid one to behold OceanOpportunity is a splendid one to behold OceanOpportunity's Avatar
    Join Date
    Nov 2006
    Location
    Providence, RI USA
    Posts
    1,378
    homebuild/custom

    pulse oximetry

    anyone know of actual research done looking at correlations of inspired PO2 and pulse oximetry (via ear or finger probe)

  2. #2
    New Member nurseling is an unknown quantity at this point nurseling's Avatar
    Join Date
    Jan 2006
    Location
    pensacola, FL
    Posts
    33

    Re: pulse oximetry

    In my experience as an intensive care nurse pulse oximetry Would have very limited use healthy lungs on air give a reading of 98 to 100%. 100% being the highest they read.

  3. #3
    RBW Member nigelh has much to be proud of nigelh has much to be proud of nigelh has much to be proud of nigelh has much to be proud of nigelh has much to be proud of nigelh has much to be proud of nigelh has much to be proud of nigelh has much to be proud of nigelh has much to be proud of nigelh has much to be proud of nigelh has much to be proud of nigelh's Avatar
    Join Date
    Mar 2005
    Posts
    1,382

    Re: pulse oximetry

    Quote Originally Posted by OceanOpportunity  View Original Post
    anyone know of actual research done looking at correlations of inspired PO2 and pulse oximetry (via ear or finger probe)
    It's only looking at the colour of the hemoglobin in the blood and a few days in a cardiac ward taught me that it read 98% for me and everybody.

    Breathing more than 0.21bar ppO2 it's probably jammed against the end stop with enough dissolved O2 that the circulation probably never needs the hemoglobin transport mechanism. It might tell you you'e about to go hypoxic but hyperoxic is the real problem.

  4. #4
    Flaccid Member abowie is a splendid one to behold abowie is a splendid one to behold abowie is a splendid one to behold abowie is a splendid one to behold abowie is a splendid one to behold abowie is a splendid one to behold abowie is a splendid one to behold abowie is a splendid one to behold abowie is a splendid one to behold abowie is a splendid one to behold abowie is a splendid one to behold abowie's Avatar
    Join Date
    May 2005
    Location
    Land of Oz
    Posts
    801
    Dammit I bought another KISS

    CK, SK, Satori, ManBagOD

    Re: pulse oximetry

    Quote Originally Posted by OceanOpportunity  View Original Post
    anyone know of actual research done looking at correlations of inspired PO2 and pulse oximetry (via ear or finger probe)
    If what you're asking about is the corelation between actual arterial PO2 and oximetry then the answer is that there is surprisingly little data.

    What do you actually want to know?

  5. #5
    hell is in the details Marc T is just really nice Marc T is just really nice Marc T is just really nice Marc T is just really nice Marc T is just really nice Marc T is just really nice Marc T is just really nice Marc T is just really nice Marc T is just really nice Marc T is just really nice Marc T is just really nice Marc T's Avatar
    Join Date
    Sep 2005
    Location
    FRANCE Carcassonne
    Posts
    836
    JOKI/rEVO/HOME MADE

    EDO04 Ray

    Re: pulse oximetry

    Quote Originally Posted by nigelh  View Original Post
    It's only looking at the colour of the hemoglobin in the blood and a few days in a cardiac ward taught me that it read 98% for me and everybody.

    Breathing more than 0.21bar ppO2 it's probably jammed against the end stop with enough dissolved O2 that the circulation probably never needs the hemoglobin transport mechanism. It might tell you you'e about to go hypoxic but hyperoxic is the real problem.
    Thinking out loud, so in hyperbaric situation if you get an O2 pulse you are close to hypoxia, interesting, , its a shame not to have a similar method for CO2, perhaps easiest to monitor that than CO2 in the loop...
    An other problem would be where to place the sensor during a dive

  6. #6
    Bubbless Box of Death Genesis is a splendid one to behold Genesis is a splendid one to behold Genesis is a splendid one to behold Genesis is a splendid one to behold Genesis is a splendid one to behold Genesis is a splendid one to behold Genesis is a splendid one to behold Genesis is a splendid one to behold Genesis is a splendid one to behold Genesis is a splendid one to behold Genesis is a splendid one to behold Genesis's Avatar
    Join Date
    Oct 2005
    Location
    Sunny Florida
    Posts
    1,462

    Re: pulse oximetry

    Quote Originally Posted by nigelh  View Original Post
    Breathing more than 0.21bar ppO2 it's probably jammed against the end stop with enough dissolved O2 that the circulation probably never needs the hemoglobin transport mechanism.
    Uh, no.

    Dissolved oxygen (in the plasma) is not sufficient on its own until you get up somewhere around 3.0 PO2, and by that point you've probably toxed!

    Hemoglobin transport is one of those amazing physiological things without which we'd all be dead......

  7. #7
    Pedant dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner's Avatar
    Join Date
    Mar 2005
    Location
    Adelaide, Australia
    Posts
    258

    Re: pulse oximetry

    A pulse oximeter measures the ratio of oxygenated to total haemoglobin. So when it says 98% it means that 98% of the total (arteriolar) haemoglobin has oxygen bound to it.

    Haemoglobin binds oxygen very well and is well saturated even at surprisingly low partial pressures of oxygen.

    In rebreather use, where we use FIO2's much higher than atmosheric air, our haemoglobin is 100% saturated with oxygen all the time.

    A fall in oxygen saturation would be a very late sign of loop hypoxia (only seen when the FIO2 falls below about 0.15 ATA), and therefore of very little use.

    Dave T

  8. #8
    Morgan's Mum bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic's Avatar
    Join Date
    Apr 2005
    Location
    Brisbane Australia
    Posts
    342
    Classic Kiss

    Sport Kiss

    Re: pulse oximetry

    Quote Originally Posted by dteubner  View Original Post
    A pulse oximeter measures the ratio of oxygenated to total haemoglobin. So when it says 98% it means that 98% of the total (arteriolar) haemoglobin has oxygen bound to it.

    Haemoglobin binds oxygen very well and is well saturated even at surprisingly low partial pressures of oxygen.

    In rebreather use, where we use FIO2's much higher than atmosheric air, our haemoglobin is 100% saturated with oxygen all the time.

    A fall in oxygen saturation would be a very late sign of loop hypoxia (only seen when the FIO2 falls below about 0.15 ATA), and therefore of very little use.

    Dave T
    sorry to the OP to go off-topic

    Dave
    On one of my trips to the chamber, i was sitting at the emergency ward and my %oxygenation was 93%. This caused some concern in relation to a possible lung perforation, so i was then subjected to chest x-ray. However the oximeter was placed on my finger that had a not-quite-recently blackened fingernail (when it met hammer). could that have interfered? Just wondering really.

    rachel

  9. #9
    Pedant dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner is a name known to all dteubner's Avatar
    Join Date
    Mar 2005
    Location
    Adelaide, Australia
    Posts
    258

    Re: pulse oximetry

    Quote Originally Posted by bendomatic  View Original Post
    sorry to the OP to go off-topic

    Dave
    On one of my trips to the chamber, i was sitting at the emergency ward and my %oxygenation was 93%. This caused some concern in relation to a possible lung perforation, so i was then subjected to chest x-ray. However the oximeter was placed on my finger that had a not-quite-recently blackened fingernail (when it met hammer). could that have interfered? Just wondering really.

    rachel
    Hi Rachel,

    Short answer - yes.

    Longer answer - the way the oximeter works is it shines two lights through the finger, one red and one near infrared. The amount of light which goes through the finger is detected by sensors. Oxygenated and deoxygenated haemoglobin absorb the lights differently and the ratio of the signal from the two sensors is used to determine what the oxygen saturation. The machine is clever in that it only uses the bit of the signal whcih fluctuates with time (i.e the pulseatile bit) to eliminate the absorption from skin, bone, venous blood etc.

    Anything that can absorb the light can potentially interfere with the signal and so maybe affect results. This might include nail polish, bruise under the nail etc. The other point is that, because the machine relies on pulsatile flow, anything which reduces blood flow to the fingers (cold, anxiety, low blood pressure) will make the reading less accurate.

    If you were on oxygen and the machine was saying 93%, then it almost certainly was an error, assuming your lungs were OK.

    Dave

  10. #10
    Morgan's Mum bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic is just really nice bendomatic's Avatar
    Join Date
    Apr 2005
    Location
    Brisbane Australia
    Posts
    342
    Classic Kiss

    Sport Kiss

    Re: pulse oximetry

    no problems with the lungs which is why i've wondered. but at least it got the staff treating me a little more seriously. ;)

    also thanks for the long answer.